Title: POC Leadership Forum
1POC Leadership Forum
- Update on the Regulations
- CLIA, CAP, JCAHO and COLA
- Ronald H. Laessig, Ph.D.
- University of Wisconsin
- Madison, Wisconsin
www.cms.hhs.gov/clia
2Purpose of U.S. Lab Regulations
- quality standards for all laboratory testing to
ensure - accuracy
- reliability
- timeliness
- of patient test results regardless of where the
test is performed
www.cms.hhs.gov/clia
3Series of six slides from Glamour Magazine
- The latest information on test quality
- Believe it or not
- 1
www.cms.hhs.gov/clia
4Series of six slides from Glamour Magazine
- The latest information on test quality
- Believe it or not
- 2
www.cms.hhs.gov/clia
5Series of six slides from Glamour Magazine
- The latest information on test quality
- Believe it or not
- 3
www.cms.hhs.gov/clia
6Series of six slides from Glamour Magazine
- The latest information on test quality
- Believe it or not
- 4
www.cms.hhs.gov/clia
7Series of six slides from Glamour Magazine
- The latest information on test quality
- Believe it or not
- 5
www.cms.hhs.gov/clia
8Series of six slides from Glamour Magazine
- The latest information on test quality
- Believe it or not
- 6
www.cms.hhs.gov/clia
9Quality POCT What drives the system?
- According to Glamour Magazine
- .(we always go right to the source)
- If Regulations are not followed...
- ...Human error follows
- J. Yost, the CLIA Lady
10Manufacturers Point of View
- Evolution in Assured Quality
- Reduction in size
- Operator ID / Patient ID
- Reagent viability
- Lock-out quality control
- Connectivity
- Data management
Roche Diagnostics
11Regulators Point of View
- U.S. Regulations CLIA
- C Clinical
- L Laboratory
- I Improvement
- A Amendments
- Regulates 180,000 U.S. test sites
http//www.cms.hhs.gov/clia/
12Regulators Point of View
- CLIA88
- (February 28, 1992)
- CLIA03
- (January 24, 2003)
13Linking Quality and Regulations
- CLIAs working premise
- CLIAs mandated quality practices will improve
quality of test results
14CLIA Regulations and the Director
- The laboratory director is responsible for the
overall operation and administration of the
laboratory the testing systems used must
provide quality laboratory services for all
aspects of test performance including the
pre-analytic, analytic, and post-analytic phases
of testing
15Quality POCT Challenge March 2004
- Dilemma for today
- With the manufacturer building-in/assuring
quality, - Where is the professional input from the
laboratory director? - Does the director just buy the technology and use
it?
16What do we know today about CLIA and Quality
Requirements?
- CLIA03 takes a Quality Systems approach like ISO
documents - The quality requirements follow the route of the
specimen - pre-analytical
- analytical
- post-analytical issues
-
17CLIA and Quality Requirements?
- The old CLIA covered the analytical requirements
in one section and quality assurance in another. - The new CLIA groups the quality requirements
under a new term Quality Assessment - Quality Assessment Quality Control Quality
Assurance - CLIA wants to consider the testing process start
to finish-hence pre, post and analytical
assessment -
18CLIA 03 and QC?
- Control procedures must (493.1256)
- Monitor the accuracy and precision of the
complete analytical process - Detect immediate errors due to test system
failure, adverse environmental conditions, and
operator performance - Monitor over time precision and accuracy that may
be influenced by changes in test system,
environment and operator performance
19CLIA03 and QC?
- Control procedures must (493.1256)
- Monitor complete analytical process
- Detect immediate errors
- Monitor over time precision and accuracy
- Conceptually, meeting the requirements sounds
easy, - But ---
20CLIA03 and QC?
- Control procedures must (493.1256)
- Monitor complete analytical process
- Detect immediate errors
- Monitor over time precision and accuracy
- Conceptually, meeting the requirements sounds
easy, - But ---
21CLIA 03 and Q C
- The Devil is in the Details
22CLIA03 The details!
- What about
- Electronic QC?
- Process controls?
- Internal controls?
- Do these
- evaluate complete analytical process
- detect immediate errors
- monitor accuracy and precision over time
- none of the above
23Details - 12 January 2004
http//www.cms.hhs.gov/clia/appendc.asp
24New Concept for QC?
- EQC
- Equivalent Quality Control
25New Concept for QC?
26What is Equivalent QC (EQC)?
- For instruments with procedural/internal
controls, test sites can choose to -
- Analyze 2 external, liquid controls per day
- OR
- Qualify the procedural/internal controls as one
of three EQC options
27EQC Option 1
- Test Systems with internal/procedural control(s)
that monitor the entire analytic process - Test site may use EQC, if test systems stability
is demonstrated
28EQC 1 - Evaluation of Stability
- Along with internal/ process controls, run 2
external QC daily for 10 consecutive days (why
10?) - If internal/process and external control results
are acceptable (what is acceptable?) - reduce external QC from daily to once / month
unless manufacturer requires more - Lab must perform/monitor the internal control(s)
in accordance with manufacturer instructions
29Option 1
- Instrument with manufacturer supplied on board
controls, run automatically.
30EQC Option 2
- Test Systems with internal/procedural control(s)
that monitor a portion of the analytic process - Test site may use EQC, if test systems stability
is demonstrated
31Option 2 - Evaluation of Stability
- Along with internal process controls, run 2
external QC daily for 30 consecutive days (why
30?) - If internal and external control results are
acceptable (what is acceptable?) - reduce external QC from daily to once / week
unless manufacturer requires more -
- Lab must perform/monitor the internal control(s)
in accordance with manufacturer instructions
32Option 2
- Hand held instrument currently advertized as
having on board, electronic controls.
33EQC 3
- Test Systems without Internal/Procedural
Control(s) - Advancements in technology have produced systems
- Capable of maintaining stable performance
specifications over time - Minimally influenced by adverse environmental
conditions and operator variance. - Test site may use EQC, if test systems stability
is demonstrated
34Option 3 - Evaluation of Stability
- Along with internal process controls, run 2
external QC daily for 60 consecutive days (why
60?) - If internal and external control results are
acceptable (what is acceptable?) - reduce testing external QC from daily to once /
week unless manufacturer requires more -
- Lab must perform/monitor the internal control(s)
in accordance with manufacturer instructions
35Option 3?
?
What belongs in this picture?
36Option 3?
Every Instrument! Vitros Hitachi aca, SMAC, etc.
Prove stability for 60 days, run external QC once
per week
37CLIA Equivalent QC (EQC)
- The director must consider the laboratorys
clinical and legal responsibility for providing
accurate and reliable patient test results versus
the cost implications of reducing the QC testing
frequency. - EQC is a choice!
38Uncle Sam
- Has Uncle Sam gone Crazy?
39Uncle Sam
- Has Uncle Sam gone Crazy?
- Maybe or Maybe Not!
40March 2004
- Does EQC and the U.S. regulations open a new era
for - new technologies
- or
- does EQC just cause confusion?
41Quality POCT What Drives the System?
- To me, CLIA QC regulations say
- Continue the old way 2 external QC/day
- Qualify the test system under EQC options and
reduce frequency of external QC
42Quality POCT What Drives the System?
- To me, CLIA QC regulations say
- Continue the old way 2 external QC/day
- Qualify the test system under EQC options and
reduce frequency of external QC - EQC Option 3 ( for a test system with no
internal/process controls) - Do you want your testing done on a system that
qualifies under this option?
43The Devil is in the details
- EQC Qualified System External QC analyzed once
per month or once per week. - EQC-1 EQC-23
44The Devil is in the Details
493.1282 Standard Corrective actions if
external QC resultsfail to meet the
laboratory's established criteria for
acceptability, all patient test results
obtainedsince the last acceptable test run must
be evaluated to determine if patient test results
have been adversely affected
45Quality POCT Drivers?
- Manufacturers Building the perfect instrument
to meet the needs of the testing environment
46CLIAs new QC/EQC Requirements
- Have a two-year phase in
- Inspection
- Citations
- NO punitive activities
47Manufacturers have EQOptions 1, 2, 3
- To be competitive in the marketplace, options 2
and 3 DONT count
48The bottom line..the Regs
- We are in a two year phase in
- We are all trying to figure out what they
mean - At the end of the day, the laboratory director
must make the call -- QC or EQC (1,2,or 3) - We are after all in the patient care business
- Legally, Morally, ethically, technically, the
patients (must) always come first.
49CLIA Regulations and the Director
- The laboratory director is responsible for the
overall operation and administration of the
laboratory the testing systems used must
provide quality laboratory services for all
aspects of test performance including the
pre-analytic, analytic, and post-analytic phases
of testing
50..and
- Speaking of Lab Directors,
- Heerrrrrs Fritz !!!!!